Does anyone have experience with a Steris/Amsco 444 not getting to temp during dry cycle? I think set point is around 240F but I'm only getting ~180F.

Thanks...


   RE: Washer/Decontaminator Report Post
Thu Oct 06 2011Reply from Dean
Dean

Try checking the dryer temp RTD that's in the ductwork just below the blower motor. you can insert a calibrated bead probe in through the screw hole that's just above it, remove the screw and place the probe in through the hole. DO NOT let the disc/shaft that the screw is holding rotate! This is easily accessible from the clean side of a double-door unit once you remove the upper panel above the door. Check the temp you get during a dry cycle on your calibrated probe against what the RTD is reading. It could be a simple matter of calibration or a bad RTD.


Dean Stephens EET, CBET, CRES
Senior Biomedical Equipment Tech
Intermediate Imaging Equipment Tech
(Not my first rodeo...)

Report Post
Thu Oct 06 2011Reply from Tool
avatar placemark

Thanks for the response. I will try that but the reason I ask is that the instruments are coming out wet and not very hot. I increased the dry time by 6 minutes and that helped a "little" but it's not optimal....