Thu Oct 13 2011

lmitchell

Baxter infusion pump, model 6201, has error message of F 38.  How can I trouble shoot to correct this error?

 


   RE: Infusion Pump Report Post
Thu Oct 13 2011Reply from GaryBiomed
GaryBiomed

Here's the link to the service manual

http://photos.medwrench.com/manuals/3191-2396.pdf

 

but you should contact AIV for support

www.aiv-inc.com
888.587.6759


Report Post
Tue Oct 18 2011Reply from Cbet_Matt
Cbet_Matt

Do you have the service manual? If not email me cbet_us@yahoo.com I'll send you a copy. F38 Manluction in the tube misloading circuit. The manual says

Check the connector CN308. Clean and or repair as needed. If CN308 is not the problem.  Replace FSR

If any of the sensors are bad the pump is down for the count unless you can find them from a second or third source or take the parts from a pump that is already dead by another problem.

Did the pump have fliud in the sensor area? "Carefully" Try to clean sensors with distilled water on a damp (not dripping wet) swab. And dry before powering up.

 



Matt

 
Report Post