Baxter infusion pump, model 6201, has error message of F 38. How can I trouble shoot to correct this error?
Baxter infusion pump, model 6201, has error message of F 38. How can I trouble shoot to correct this error?
Here's the link to the service manual
http://photos.medwrench.com/manuals/3191-2396.pdf
but you should contact AIV for support
www.aiv-inc.com
888.587.6759
Do you have the service manual? If not email me cbet_us@yahoo.com I'll send you a copy. F38 Manluction in the tube misloading circuit. The manual says
Check the connector CN308. Clean and or repair as needed. If CN308 is not the problem. Replace FSR
If any of the sensors are bad the pump is down for the count unless you can find them from a second or third source or take the parts from a pump that is already dead by another problem.
Did the pump have fliud in the sensor area? "Carefully" Try to clean sensors with distilled water on a damp (not dripping wet) swab. And dry before powering up.