caliberation password

Hi

I have an Ohmeda ADU plus anesthesia work station, it went into selftest, and it remains in this state, the service manuals said that insipratory valve is the problem , I need to calibrate it, password is mandatory, any body have it, please


   RE: Anesthesia Gas Machine Report Post
Mon Mar 09 2015Reply from jose maria
avatar placemark

Hi there, I have the same problem, the machine doesn´t pass the selftest.

 

Did you find the Calibration and Diagnostic Passwords???


Report Post
Mon Mar 09 2015Reply from Cbracci
avatar placemark

Install/Service   46-1-47

Service      30-19-7

Calibration    67-57-21


Report Post
Mon Mar 09 2015Reply from saadibrahim
avatar placemark

Thank you very very much

I will try it and I hope it work

thanks again


Report Post
Tue Mar 10 2015Reply from jose maria
avatar placemark
Hi thanks the password worked!. I went to the A-EV1 diagnostic page and the only message that came was:   Range Flow Error.   and these:   start: insp cal   start: insp ver   Start: backup   start: Exp Valv   Start Flw zero   Start: flw sens   Start: pilot pr   start peep val       The only error was the Flow range error.  I think it might be the inspiratory valve

Report Post
Fri Apr 28 2017Reply from Rosy Juarez
avatar placemark

Thank you very much for the passwords!


Report Post