Hi everybody, I have a problem with a Siremobil equipment. I have a 4028 error code and 7150 error code. Please help me ando if somebody have the service manual please send it to llilli_68@hotmail.com thanks team.
error codes and service manual
=======
Error E4028 is usually a bad Memoskop.
--> Contact us for a replacement quotation (exchange, or repair & return) wayne@columbiaimaging.com
E4028 Factory Help => Memoskop unexpected acq end signal
Possible trouble:
1. Host registered unexpected acq end signal of the Memoskop.
Possible help:
1. Check faults on the Memoskop acquisition cable (D1.X10.27).
2. Check Memoskop.
3. Check connection cables to the Memoskop.
-------
Error E7150 is usually stripped teeth on the ring gears for the slot diaphragm.
--> Contact us for a replacement quotation (exchange, or repair & return) wayne@columbiaimaging.com
E7150 Factory Help => Slot diaphragm rotation index not found
Possible trouble:
1. The slot diaphragm rotation index position was not found
during initialization.
Possible help:
1. Check the index switch
2. Check the index switch wires.
3. Check the motor.
4. Check the motor wires.
5. Check the mechanical movement.
6. Board D1 defective
Wayne Horsman
Columbia Imaging
wayne@columbiaimaging.com