New prospective customer says his 414 Podiatry chair not responding to footpedal commands, Anyone got a wiring diagram for the foot control? Would be much appreciated,
Sun Oct 23 2011
Mike@bryant's Electronic Servicing & Testing
Mon Oct 24 2011Reply from Dean

I just uploaded the service manual to http://photos.medwrench.com/manuals/3697-2868.pdf
If this is the chair that I am thinking of, pay particular attention to the dip switches on the main control board; something as simple as the wrong line frequency being selected can cause lots of gremlins with these.
Hope this helps...
Dean Stephens EET, CBET, CRES
Senior Biomedical Equipment Tech
Intermediate Imaging Equipment Tech
(Not my first rodeo...)
Report Post
Fri Oct 28 2011Reply from rdttech

MIDMARK.com/ tech library/medical/power tables/400 series/414!
If you have questions just ask!
Report Post
Sun Nov 20 2011Reply from Frank Foo
Do you have directions for the foot pedal controls to reset it. I had it years ago and lost it. At that time when I reset the foot pedal the chair worked again.
Thanks
Report Post
Sun Nov 20 2011Reply from Frank Foo
Mon Nov 21 2011Reply from rdttech