Testing of Pulsed Output

Does any have a service manual or the procedure to test the pulsed output of the Sabre Genesis?

Boyd Campbell, CBET CRES

Southeastern Biomedical


   RE: Electrosurgical Unit (ESU) Report Post
Tue Oct 14 2014Reply from jmandell
avatar placemark

Unfortunately not all ESU analyzers can verify Pulsed Waveforms. In my experiance, I've used the Fluke QA-ESII ESU Analyzer.

This are the procedures that I created for testing the Conmed 5000, which is closely related to the Saber Genesis.

Prepare ESU by pressing the square sine wave icon while in the Standard COAG mode and set power to 10 watts.

  • To test, Press F3– START on ESU Analyzer first.
  • Next press blue COAG key on monopolar handswitch.
  • Press F3– STOP when power is displayed on ESU Analyzer.
  • Release blue COAGkey on monopolar

 

 


Report Post
Tue Oct 14 2014Reply from Boyd Campbell CBET, CRES
Boyd Campbell CBET, CRES

Thanks!