How to silence alarms under specific circumstances.

I am in a pain clinic where all cases are MAC.  We use the Narkomed 2b as a back-up in case of emergency, but 99% of the time only utilize the aux. O2.  However the machine is turned on and checked daily.  How do you silence the alarm that goes constanly saying %02Low, etc.  You can silence for 120 seconds, but it is annoying to have to press the silence alarm button every 2 min during a 10 hr day!!  What is the trick to silencing that alarm for the duration, yet keeping the machine ready to go as needed?


   RE: Anesthesia Gas Machine Report Post
Mon Oct 20 2014Reply from Rix
avatar placemark

Nominal O2 concentration in room air is 21%. When you turn on the 2B, the default low O2 alarm setting is 30%. Select the low O2 alarm setting and lower it with the down arrow as far as you can (usually 18%). That should eliminate your alarms until the unit is switched on again , where this process will need to be repeated.


Report Post
Wed Oct 29 2014Reply from BrentS
BrentS

The Narkomed family resets the low O2 limit alarm level everytime you turn the machine on to 30%. You will have to manually change it each time you turn on the machine.


Brent SilviusAnesthesia TechnicianBiomedical Sales and ServicesMercuryMed.com
Report Post
Thu May 25 2017Reply from BrentS
BrentS

Turn down the Oxygen low alarm level to either 18 or 21%


Brent SilviusAnesthesia TechnicianBiomedical Sales and ServicesMercuryMed.com