Arizant Bear Paws Service Procedures
PM Process:
- Perform electrical safety procedure.
- Install temperature test apparatus on end of hose.
- Expose test pots by lowering filter shroud.
- Replace filter.
- Extend hose to full length.
- Turn unit on to full value of thermostat control.
- Allow 10 minutes to stabilize
- Take reading (43.0C +/- 0.5)(if out, see calibration below).
- Push over-temp switch to advance temperature uncontrolled.
- Alarm lamp should light below 56C (There is no control of this value).
Calibration Process:
- Perform PM procedure until step 6 completed.
- Using small flat screwdriver, adjust temp reading gradualy through adjust hole under shroud.
- When 43.0C is met and stabilized, proceed with over-temp steps 7 & 8 above.