GE AMX 4 Multiple Errors

Unit Fails Test 4, Gives Check Sum Errors 6, 7, 10, 11.

Tried to calibrate Generator, but whenever any calibration is performed unit presents with Ram Readback Error when values are input.


   RE: Mobile Radiography Report Post
Fri Nov 11 2011Reply from JC Killinger
JC Killinger

See posting http://www.medwrench.com/?community.posts/threadNo/4953/subject/Error-code/

In these cases check power/batteries then factory reset is usually the best bet.


Report Post
Mon Sep 16 2013Reply from Wkrach
avatar placemark

Make sure you are exiting the calibration properly. If you don't it will corrupt the cal data.


Report Post