Noisy fan motor

Noisy fan motor, noise is abated with slight clockwise pressure to fan motor (when facing unit faceplate). However, noise returns as unit warms up. Any suggestions?


   RE: Hot/Cold Therapy Systems Report Post
Tue Nov 15 2011Reply from GaryBiomed
GaryBiomed

Probably the propeller in the basin rubbing against the plastic cage and hose assy.  open up the basin & check the propeller.


Report Post
Tue Nov 15 2011Reply from rdttech
rdttech

Typically a noisy or slow moving pump would cause the overheat to trip; are you having this problem?

As Gary suggested check impeller; if no crack or visible wear the bushings are worn on the motor; replace motor!


Report Post
Wed Nov 16 2011Reply from sulsapj
sulsapj

Thanks, it was the impeller hitting the pump housing. This pump has taken a couple of falls and the motor mounts are a bit askew. No matter what I tried short of replacing the tray would remedy the problem so I ended taking a Dremel to the impeller vanes and slightly shortening them equally.


Rick Trexler, RVT, CBET
Report Post
Wed Nov 16 2011Reply from sulsapj
sulsapj

No overheating issues. See reply to Gary's post. Thanks for the input.


Rick Trexler, RVT, CBET