Errror Code F06 and 250.0

Does any one know error codes F06 and 250.0?

Thanks


   RE: Contrast Media-Injector System Report Post
Mon Nov 28 2011Reply from Ryan C.
avatar placemark

F06 indicates an issue with the plunger potentiometer feed back.  Is the issue intermittant or steady?  Check all your cables to ensure they are all secure - especially the DB connectors on the back of the main unit.  Check the 6' head cable for damage as well as the extension cable which plugs into the head cable.  It is possible the system is out of range - to test, manually advance the plunger via the manual knob by rotating 2-3 times, then cycle power.  For more corrective steps or further service assitance, contact Ryan at 888-667-1062.


Ryan Clarke - ISS  |  We Are ISS  |  888-667-1062

Report Post
Tue Nov 29 2011Reply from sdimedical
avatar placemark

Ryan C. thanks very much for the advice, I will call you after we check the unit.

 

Robert


Report Post
Mon Dec 05 2011Reply from Deb
avatar placemark

Error code 250 F06 is a Pot Calibration data recieved from memor is invalid recalibrate the posts if the condition persists replace the HCU


Report Post
Tue Dec 11 2012Reply from samik
avatar placemark

hi guys,

hope someone can help me here,

i have envision ct injector, and it's giving error E82 on the injector head.