Blowing Gfic Rec

I have a 2540M That when turned on blows GFCIs. I checked for shorted wires and found none. Any idea as to what may cause this?


   RE: Sterilizers/Autoclaves Report Post
Wed Jul 22 2015Reply from Duraline Systems Inc.
avatar placemark

You most likely have a faulty heater. There are four of the wired in circuit which means they each have to be removed and check indvidually. If you would like further help feel free to contact our service department at 877-561-0500 or service@duralinesystems.net

Thanks,

Darren Walker

Duraline Systems Inc.

www.duralinesystems.com

 


Duraline Systems Incorporated
39 Western Hwy. West Nyack, NY 10994
www.duralinesystems.com
877-561-0500

Report Post
Wed Jul 22 2015Reply from bayside1
avatar placemark

Check the three switches under the multivalve.

They have three positions comm, no and NC.

Press down on the levers with the power off and unplugged.

Each should make a clicking noise. No click equals a bad switch. The middle one is the most problematic.

Be careful with the heaters. It's easy to screw them up when testing as advised by another company.

Bayside Technical Services

508-735-4817


Report Post
Wed Jul 22 2015Reply from Jeff_EBS
avatar placemark

I have had this problem before and it was the heaters. Be sure to check the heater individually and from each of the terminals to ground. The terminal to ground test is what was causing my problem.


Report Post
Mon Aug 03 2015Reply from jeverson
avatar placemark

Yes on the heaters! I have even seen them short ONLY when heated OR bent around the chamber!


Report Post
Mon Aug 10 2015Reply from Jeff_EBS
avatar placemark

GFCI normally don't  trip due to a short. GFCI’s look at leakage between the neutral & ground wires.


Report Post