converting sterrad insst to autoclave

Hi,

would appreciate if you can advice what should be the remedial action after converting sterrad inst to autoclaving and found that the light cable deteriorated.

Do I revert back to sterrad?


   RE: Sterilizers/Autoclaves Report Post
Mon Jul 27 2015Reply from stericert
avatar placemark

the question is what the light cord IFU's said. If they say don't autoclave, then you shouldn't. If they say to do it, then the manufacturer should be contacted. In the meantime, revert to Sterrad.


Report Post
Tue Jul 28 2015Reply from ylchan
avatar placemark

Thanks for the reply. The IFU states that it can be autoclaved. However it did not specify that changing the method of processing will cause damage to the light cable.


Report Post