error message 1,2,3,4 completed

error 45 B then image turned off


   RE: Mobile Radiography Report Post
Thu Jan 05 2012Reply from howletrc
Thu Jan 05 2012Reply from E. Smith
E. Smith

Fil/Kv Board Defective

 



Eric Smith – President [b]SOS Medical Imaging, Inc.[/b] P.O.Box 1046 Cornelius, NC 28031 704-995-6502 esmith@sosmedicalimaging.com A 2015 DOTmed Top 100 Equipment Seller [url=http://]http://www.SOSMedicalImaging.com[/url]
Report Post
Sat Jan 07 2012Reply from xrayxpert
avatar placemark

If the KV/Fil board doesn't fix your problem then it's the 60hz inverter. There should be 27Vdc across AMX1-A4-C6 when 60hz inverter is on (prep).