E-21 error code

I keep getting this error does anyone know how to repair this dont have a manual for this unit.


   RE: Centrifuge Report Post
Wed Jan 11 2012Reply from C. Mash
C. Mash

At what point does it give you the error? (Start of run - Middle - End) With a little more information, we can narrow it down to what is causing the error code.


Ozark Biomedical
Senior Technical Specialist
800-457-7576
Report Post
Wed Jan 11 2012Reply from Ed
avatar placemark

This error code (E21) refers to a missing, non-standard or incorrectly installed rotor.  Are you using a compatible rotor?  Is it correctly installed?  Was there a power failure during a run (power switch moved to OFF, power cord disconnect or, intermitten power at the outlet?)

 

If, after checking all of the above, you still have this error code, let me know.  There is one more check you can perform to help isolate the problem.

 

-Ed Blanchard, CBET

Gilbert, AZ

 

Report Post
Wed Jan 11 2012Reply from Ed
avatar placemark

BTW- here's a copy of the operator's manual:

http://www.djblabcare.co.uk/djb/data/sheet/75967273/Sorvall_Legend_RT_Plus-User_manual.pdf.

-Ed Blanchard, CBET, Gilbert, AZ

 

Report Post
Wed Jan 11 2012Reply from MedWrenchManager
MedWrenchManager

Ed,

Thanks for sharing a link to the user manual. I have also uploaded it to the product page, under the resources tab, here.

Thanks!


Report Post
Thu Jan 12 2012Reply from Emma
Emma

Hi adavila,

I recommend you contact our Technical Support team at 800-438-4851. Make sure you have the unit's serial number of the unit prior to contact so the team can verify the exact model and age.



Kind regards,
Emma

Thermo Fisher Scientific
www.thermoscientific.com

Follow @servingscience | Register www.thermoscientific.com/servingscience
 
Report Post
Wed Jul 25 2012Reply from C. Mash
C. Mash

The E-21 states it is a rotor recognition disturbance. The service manual points to the sensor wiring, we have noticed the error usually occurs from a lack of grounding. Double check all the ground connections including the one on the main pcb next to the fuses, remove it and reattach. I hope this helps and as always, thanks for using MedWrench.


Ozark Biomedical
Senior Technical Specialist
800-457-7576
Report Post