Boot Up Error 2501

Need some help with an AGFA

DX-S.

Getting 2501 error at boot up. Error log says to check cabling to the scanhead. Replaced the SHI board and the Scanhead.. Same error. Anyone have any ideas?


   RE: Computed Radiography Report Post
Wed Sep 16 2015Reply from Estrella
Estrella

Have you checked the wire to SHI board and the scanhead?
Can you attach some pictures?



Best regards,
Estrella.


If you have some technical problem with your imaging equipment, send me an email, and i will reply asap to help you.

Email: jmstreya@gmail.com
Report Post
Fri Sep 18 2015Reply from AgfaDoog
AgfaDoog

Yep, check it with a voltmeter at the connector, sometimes is just a false contact.


"...Hello, IT... Have you tried forcing an unexpected reboot?..."
The IT Crowd
Check my LinkedIn profile for more info: https://www.linkedin.com/in/aacch/?locale=en_US