Baxter 6201 constantly beeping

I found the battery fried. Replaced it and tried getting in configuration but it won't quit beeping and I can't get in configuration. Any suggestions?


   RE: Infusion Pump Report Post
Wed Jan 11 2012Reply from Dean
Dean

There is an internal clock/ROM battery, also. I have fixed this issue in the past by replacing that battery, but it only works about half the time. If replacing the internal battery doesn't fix it, it's usually a sign that the main board is fried.


Dean Stephens EET, CBET, CRES
Senior Biomedical Equipment Tech
Intermediate Imaging Equipment Tech
(Not my first rodeo...)

Report Post
Wed Jan 11 2012Reply from tjgdjn
avatar placemark

I removed that battery and tested it. It is reading 3.069v. I guess it is their board. Thanks for the advise.


Report Post
Wed May 30 2012Reply from mouser69
avatar placemark

there is a battery fuse on the harness this could also be the issue


Report Post
Wed Oct 29 2014Reply from brodak
avatar placemark

I had this problem on a 6301 and it was the 1 amp fuse (F302) on the main board.