Our monitor shows a pulse rate based on ECG...Can this be changed to have the pulse rate derived from the pulse oximeter instead?
Thank you,
Our monitor shows a pulse rate based on ECG...Can this be changed to have the pulse rate derived from the pulse oximeter instead?
Thank you,
Here's a link to the user manual
http://photos.medwrench.com/manuals/1529-1462.pdf
page 40 (1-6) talks about
(2) SpO2 Numeric. The patient's Oxygen Saturation value is displayed here if P2
is On.
(3) EtCO2/Respiration Display. (If EtCO2 Option is installed) The patients
EtCO2/Respiration values are displayed to the right of NIBP and, if P2 is On,
the SpO2 numeric. This display consists of numeric values concerning EtCO2
and Respiration plus Respiration Waveform Trace.