Alternate O2 and vap warning on system startup

After system start up there are two popup boxes located at the top of the machines checkout screen that say "Set Alt O2 Flow! Check Agent Setting!".

Other machines do not display this message and this only happens when first powered up. If you restart the machine this message does not re-appear.

Could this be aproblem with the Alt O2 switch, mixer, transducer issue or something else?

* Machine has version 8.01 software


   RE: Anesthesia Gas Machine Report Post
Tue Apr 25 2017Reply from IPET
avatar placemark

Hello ,

do you have any solution of this problem that you have described ?

 


Report Post
Wed Apr 26 2017Reply from JOE
avatar placemark
Could be a mixer problem. In your maintenance menus check the logs if there is a fault it will show up there. Also if there is a fault with the Alt O2 switch it will show up.

Report Post
Wed Apr 26 2017Reply from MedWrench MM
MedWrench MM

If you are in need of immediate assistance please feel free to reach out directly to our Anesthesia Gas Machine sponsors. They are happy to help.

Soma Technology

Denova Medical

Just be sure to tell them you were referred by MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Thu Apr 27 2017Reply from Cbracci
avatar placemark
if you have the service software, it will sometimes show up in the gas schematic or the electrical. have you tried another vap ?

Report Post
Tue Aug 27 2019Reply from BrentS
BrentS

You should first check the machine error history log for more information about your issue. The technical service manual has documentation for each of the error messages found in the history.

It could be your mixer assembly, which is typically a defective solenoid. Performing a zeroing or calibration on the mixer. You could even have a defective Alt O2 switch or harness causing your problem.


Brent SilviusAnesthesia TechnicianBiomedical Sales and ServicesMercuryMed.com
Report Post
Fri Sep 01 2023Reply from badsi
avatar placemark

Hello , I have the same problem , is there any action to do to solve this ? change mixer ? calibrate , or other , alt O2 flow tests fail (circuit checkout ) > 400 ml/min . Thank you 


Report Post
Fri Sep 01 2023Reply from Cbracci
avatar placemark

Try bleeding all the gas from the machine.  Start the machine and let it go through initial test then plug in the wall gases again. If the message clears then It could be your O2 or ALT o2 valves on the mixer board not zeroing ( check error codes ).  Cheaper to replace those.  Also try calibrating the Mixer board in the service section. Otherwise changing the Mixer board will generally fix the issue.  If none of that works then its possible for it to be the ALT O2 switch.  


Report Post