I have bad sample suction, the blood do not reachs the ref. Electrode, and we
got bad patient result, while the control results are good, some times we have
also Rinse Error. Please help
I have bad sample suction, the blood do not reachs the ref. Electrode, and we
got bad patient result, while the control results are good, some times we have
also Rinse Error. Please help
Assalam Alaikum,
Dear,
This is a common error in all ABL series, but this is because of lack of weelkly maintainence. The solution of the your question is
1. Check the level of Rinse, Cal 1 & 2, standard. If it is low replace it.
2. Protein removal. (Procedure in User Manual)
3. check for All electrodes & Controls expiries. If expired replace it.(Fallow the Instruction in manual)
4. Run Calibration 1 & 2. & Control Level 1, 2, 3, 4.
Inshallah every thing will be alright & you will never find errors in future.
Irfan,
KAMC-MOH. KSA.
follow the above mantion instruction
any radio meter cd analyzer problam contact to me
syed faiz
Bio-medical engineer
Please help with very high sensitivity status (above 500) of PCO2 electrode in ABL5 blood gas analyzer. Thank you in advance
hi, someone help me. i need operator's manual of abl5. pls send me 1 at my email
ill appreciate your help,
thank you in advance..