Radiometer ABL5

I have bad sample suction, the blood do not reachs the ref. Electrode, and we

got bad patient result, while the control results are good, some times we have

also Rinse Error.  Please help 

 


   RE: Blood Gas/pH Analyzer Report Post
Mon Jan 16 2012Reply from Irfanlab
avatar placemark

Assalam Alaikum,

Dear,

This is a common error in all ABL series, but this is because of lack of weelkly maintainence. The solution of the your question is

1. Check the level of Rinse, Cal 1 & 2, standard. If it is low replace it.

2. Protein removal. (Procedure in User Manual)

3. check for All electrodes & Controls expiries. If expired replace it.(Fallow the Instruction in manual)

4. Run Calibration 1 & 2. & Control Level 1, 2, 3, 4.

Inshallah every thing will be alright & you will never find errors in future.

Irfan,

KAMC-MOH. KSA.


Report Post
Wed Feb 08 2012Reply from bio-medical engineer
avatar placemark

follow the above mantion instruction

any radio meter cd analyzer problam contact to me

 

syed faiz

Bio-medical engineer

 


Report Post
Sun Sep 01 2013Reply from Nasr
avatar placemark

Please help with  very high sensitivity status (above 500) of PCO2 electrode in ABL5 blood gas analyzer. Thank you in advance

 


Report Post
Sat Mar 17 2018Reply from mesmesiglen
avatar placemark

hi, someone help me. i need operator's manual of abl5. pls send me 1 at my email

ifernglen2013@gmail.com

ill appreciate your help,

thank you in advance.. 