Datascope Pasport 2 Patient Monitor

I am biomed engineer and we have this monitor,  but it is not delivering NIBP. It's inflating and deflating  small air. Please tell me what is the problem. Thanks.

 


   RE: Vital Signs/Multi-Parameter Report Post
Tue Jan 24 2012Reply from GaryBiomed
GaryBiomed

Probably need to get a new NIBP module from Datascope/mindray

1-800-288-2121

P/N 0670-00-0746-01

or

P/N 0670-00-0798-01

depends on serial number


Report Post
Fri Nov 09 2012Reply from sprintracer
avatar placemark

run the NIBP module ck in the service mode. It is helpful to have the test chamber. Should run these checks during the PM. motor test, leak test, bleed rate, etc.