siemens sonoline omnia scan field problem

Hello all,

               My Siemens Sonoline Omnia developed some problem. The scan field does not come up when power on but it displays the frequency of the plugged transducer.

What could be the cause of this problem??

Thanks


   RE: Ultrasound Systems Report Post
Tue Jan 31 2012Reply from CCMS, Inc.
CCMS, Inc.

Hi Segun,

Have you tried a different probe port and/or probe? I would also check all the power supplies.

After that, re-seat the boards one by one to see if that has any effect.

From there, it is most likely a power supply that feeds one of the front end boards or the HV supply. Also, it could be a faulty front end board or the HV supply themselves.

Best regards,

Chris

C&C Medical Solutions, Inc.

_____________________

Christopher Turner

Phone: 317-219-3616 x103

Fax: 317-770-0560

Yahoo IM: caruso122

Skype IM: cturner122

Email: chris@ccmedicalsolutions.com

On-Line Store: www.ccmedicalsolutions.com



C&C Medical Solutions, Inc.

_____________________

Christopher Turner

Phone: 317-804-8099 x103

Fax: 317-804-8099

Yahoo IM: caruso122

Skype IM: cturner122

Email: chris@ccmedicalsolutions.com

On-Line Store: www.ccmedicalsolutions.com
Report Post
Sat Feb 18 2012Reply from segun_ademetan@yahoo.co.uk
avatar placemark

pls do you have these front end boards.

thanks

 


Report Post
Sat Feb 18 2012Reply from CCMS, Inc.
CCMS, Inc.

Hi Segun,

Part numbers?

Regards,

Chris



C&C Medical Solutions, Inc.

_____________________

Christopher Turner

Phone: 317-804-8099 x103

Fax: 317-804-8099

Yahoo IM: caruso122

Skype IM: cturner122

Email: chris@ccmedicalsolutions.com

On-Line Store: www.ccmedicalsolutions.com
Report Post
Mon Feb 20 2012Reply from segun_ademetan@yahoo.co.uk
avatar placemark

  13338000503413 S/N 000248    ( A 37 LNC REV 2 )
  13338991105446 S/N 000831    ( A 39 STEERER REV 1)
  13338000503423 S/N 000210    ( A 36 HARMONIZER-2 REV 3 )


Report Post
Mon Feb 20 2012Reply from CCMS, Inc.
CCMS, Inc.

Hi Segun,

Those part numbers look a little long. Please email me at chris@ccmedicalsolutions.com with clear pictures showing all markings so that I can try to indentify the orderable part number.

Regards,

Chris



C&C Medical Solutions, Inc.

_____________________

Christopher Turner

Phone: 317-804-8099 x103

Fax: 317-804-8099

Yahoo IM: caruso122

Skype IM: cturner122

Email: chris@ccmedicalsolutions.com

On-Line Store: www.ccmedicalsolutions.com
Report Post

This site uses cookies. By continuing to browse the site you are agreeing to our use of cookies. Please review our Privacy Policy for more details.
I Agree