Can anyone tell me if the transducer for FFsonic UF-4100 can be changed.
I want to be able to use a Vaginal Transducer with the machine.
Is this possible?
Is this possible?
Hello Doc Scott,
That unit was able to run the following probe:
FUT-TV35-5 TRANSVAGINAL 12MM TRANSDUCER (4.0/5.0/6.0MHZ) FOR FUKUDA UF-4100
If you have never run this probe on your specific unit, I recommend you call Fukuda and give them your unit serial number and software version to confirm your unit has the necessary option/software to run it.
Best regards,
Chris
hello sir
to connect vaginal probe you need a connector board to connect the tow probes at the same time
this connector board connects inside the machine on the 4 sockets j32,j33,j34,j35
convex probe connects to the board and there is another socket on the board to connect vaginal probe
so you need to buy a connector board