Tue Mar 15 2016

your help saves lives

Hi. I am an engineer from Aleppo,Syria

I was asked by the Red Crescent to repair thier only BTS-310 device. The device is working all good but the lamp is not working. I replaced the lamp but still no light..

The lamp is labled as 6 volts but the terminals of the lamp socket gives 25 volts on my voltmeter. What could be the reason the lamp is not working. I know that the device should be sent to a service centre but war in Syria did not agree. I am trying to repair it under bombing .. hoping it could help save some lives. if any one has any circuitry schematic or any thing that could help.


   RE: Clinical Chemistry Analyzer Report Post
Thu Mar 24 2016Reply from pirro
avatar placemark

Hi,

The lamp is 12V 20W, you shuold find a lamp with clear glass(no UV filter) and with large working hours.

 

Pirro


Report Post