Sun Feb 19 2012
How to Print an ECG report in two pages
I have been struggling to set my MAC 5000 print an ECG report, six strips per page i.e lead I, II, III, AVR, AVL and AVF in page one and then lead V1-V6 in page 2. Kindly help. Currently the writer print all leads in a single sheet.
Thu Feb 23 2012Reply from Statbiomed
Manual is available online via pdf . If you need it send your email address
Fri Feb 24 2012Reply from GaryBiomed
GaryBiomed
glancing over the user and service manual on the GE web site, it doesn't look like it has that feature.
unless there is a setting in the system setup menu, which you need a password for.
GE web site for manuals
http://www.gehealthcare.com/usen/service/biomed_tech_selfservice/services_user_doc/products/support.html
Fri Feb 24 2012Reply from ozzie
Thanks very much for the comments
Thu Mar 07 2013Reply from DJoy
You can call GE applications support and if the Mac 5K can do it they should be able to walk you through the set up procedure.
David A. Joy
DMD Biomedical Inc.
d.joy@dmdbiomedical.com