Ecg data from the Case 8000 is fat-lined and has artifact. The Case 8000 it's self acquires and displays clear ECG. The interconnect cable between the vivid ultrasound and Case 8000 has been replaced and the connectors appear to be in good shape. Does anyone have experience with this?


   RE: Electrocardiograph (EKG / ECG) Report Post
Wed Apr 13 2016Reply from Amscopolaris
Amscopolaris

first check the ECG electrodes system, not quality electrodes makes noise

above all electrical It occurred in the electrical system (Earth - phase-neutral) Harmonic fluctuations makes noise