FU/CCI ERROR 70

FU/CCI ERROR 70 What does it means? What can ? do? please help me!

 


   RE: Centrifuge Report Post
Sun Jul 10 2016Reply from AussieAndy
avatar placemark
CONTROL - ERROR 07 Error Lid lock message is faulty. Control panel indicates: lid lock closed Supply board indicates: lid lock open Consequences of the error The centrifuge slows down until the “open the lid” release occurs. Cooling system is switched off. Cause of error • Flat ribbon cable to the control panel (A5) is defective • Supply board (A1) or control panel (A5) is defective Delete error code Carry out a MAINS RESET.

Report Post
Wed Dec 14 2016Reply from Hettich Tech Service - USA
avatar placemark

Hello,

Has this problem been resolved?

Is the centrifuge a Roto Silent 630 or Roto Silent 63?

Could you please reverify the error as there is no error FU/CCI 70 in the repair manual. Thanks!



Best Regards,

Darrall Pride

Hettich Instruments

Beverly, MA USA