After power on, displayed two error messages: Error 200 - "Defective energy shutdown"... Error 203 "Defective pulse shutdown". I can't find the error list or service manual. Can You explain me what I must check? And what can do with this problem with nyself.
Errors 200 & 203
Facing the same issues. After turning on 200 ,or 203 error appears on screen.
Please let me know the Solution
Report Post
error 203 is when the cavity is at the end of life, check the hours of the laser and the shots , if you have many you have to change the cavity
Regards
Report Post
Thanks for valuable information. and glad to know reason
Following message appeared on screen
Error 200
Defective energy shutdown.
After turning off and on again and again (3 times). It becomes ready...
Please help solve the problems
Some times error 203 that was already replied..
Thanking you
Shashikant shinde
Report Post
Hi,
Pusch INFO KEY and let me know laser hours, you can measure the otput power with power meter, swicht on the Yag laser select max power mode 1 , measure the power , if the power under 8 mJ you need change the cavity
Regards
Report Post
Thanks for replay
Working time 413 hours.
Version 2.1.3
Please let me know how to change the cavity.
Its location and place.also how to upgrade the version.
Please help me solve this issue
Thanks
Shashikant Shinde
Report Post
You need new cavity, special tool to adjust, PCMACIA with software version 2.17
and power meter
i am from Spain
Report Post
Pivotal guidance is helpful to solve the issue.
If any service support is required then please guide us.
Thanks
Shashikant Shinde