trackball-GE Ultrasound - LogiQ9

Does anyone have any procedure for replacing the trackball on the LogiQ9?

I have a bad one and the unit is just out of warranty. I can't find any information about this repair! Any help will be appreciated!


   RE: Ultrasound Systems Report Post
Wed Mar 14 2012Reply from dferguson
Wed Mar 14 2012Reply from dferguson
Wed Mar 14 2012Reply from rsiefers
avatar placemark

I checked there, no luck

service manual doesn't show trackball replacement

thanks anyway!


Report Post
Wed Mar 14 2012Reply from dferguson
dferguson

If you are unable to find a full removal and replacement procedure, then check for the procedure for the next higher assembly (i.e. lower control panel).  After that there are only two screws and two cables to remove and/or replace.


 
 

Report Post
Mon Apr 09 2012Reply from Radiology OneSource
Radiology OneSource

Which BT0 system do you have?


Debbie Cole
Radiology OneSource Inc.
Report Post
Wed May 02 2012Reply from rgashi
avatar placemark

remove touch panel with 4 screws through back, remove ALL screws under the lower user interface/control panel. now you'll see the trackball under neath the lower user panel two screws and it comes off.