With the lockouts off, whenever you press a function on the patients control board, all functions lockout. Any ideas what causes this? Thanks
Mon Mar 12 2012
Bed Lockout get set by patient trying to use their position buttons
Tue Mar 13 2012Reply from Todd S

Crack open the siderail. There is a chip that is likely popped out partially on the board. If you have a spare board you can see where the chip is and poss press on the outside of the siderail to re-seat.