Mon Mar 12 2012

Bed Lockout get set by patient trying to use their position buttons

With the lockouts off, whenever you press a function on the patients control board, all functions lockout. Any ideas what causes this? Thanks


   RE: Bed Report Post
Tue Mar 13 2012Reply from Todd S
Todd S

Crack open the siderail. There is a chip that is likely popped out partially on the board. If you have a spare board you can see where the chip is and poss press on the outside of the siderail to re-seat.


Report Post
Thu Apr 11 2013Reply from Weso
avatar placemark

Just realized I never said thank you. That was the problem.


Report Post
Fri Apr 12 2013Reply from Todd S
Todd S

You are very welcome. Glad it helped 


Report Post