Both E13,E77 displayed on the panel.
Now, I can't open the door.
Pressure in chamber: 155-160 mbar, so the door can't open.
I dont know how to open the door.
Both E13,E77 displayed on the panel.
Now, I can't open the door.
Pressure in chamber: 155-160 mbar, so the door can't open.
I dont know how to open the door.
To open the door with a E code you need to perform a ''clear silent abort'' With the power on you have to flip the number 3 and 7 dip switch on dip sitch SW1. But be carefull never flip the number 2 and 7 (reset site set up) because you will lost the programation. So when you flip the 3 and 7 dip swith on SW1(dip switch) you wait 2 minutes
, the sterilizer will release the pressure after about 2 to 3 minutes you have to power off the sterilizer, you flip back to open the dip switch 3 with the number 7 switch (close position) and after you power on the sterilizer this will perform a ''reset cycle to idle'', you go to the dispaly and this wrote ''enter master code'', you enter '' 3 stop 33 ', as code. you cobe back to the back of the sterilizer, you power off the sterilizer you flip back to open postition the number 7 dip switch of SWI and you power on the sterilizer. On the display you should have only a C code or not. you start the sterilizer normaly so the C code will desapear and the cycle will start.
Richard
CHRDL Hospital
Do you have all these C and E codes problem on the same sterilizer?
If positive, check the 24 volts and 5 volts power supplies. When the powers supplies are defect they made weirds problems. But 2 year ago one of our sterilizer did weirds problems and the only way to fix it was to replace the cpu board.
Richard
electronic technician
CHRDL hospital
Ok.
Sterivac Machine operated normal.
I used function Vent Solenoid..
Thanks for support always.