machine stops in the middle of exposure, gave error 715/080.
Siemens Multix Pro Top
Error 715/80
Meaning:
· Minus KVmin error the negative kV (kV +) has more than 10 kV difference .
· The tube, being active while this error was sent, can be read in the environment info of the errorlog since XCU VB07A.
Description:
· The minus KVmin monitoring becomes active if the -KV actual value is more than 10KV less than the -KV nominal value
· Display LED D100.V165 KV MIN-
Possible causes and action:
· Check minus KV actual value at D220.X18.A8 and D100.X61 Meßpunkt KV- (see wiring diagram X2206-15 / X2169-15) at both test points 1V = -10kV
· Check ribbon cable between D220.X1 and D100.X1
· Perform "Diagnostic/Inverter Test"
· Check oscillating current on D100. X62 I_LAST (1V = 50A) and compare with diagram ERROR 711
· Check connection screws in the intermediate circuit and inverter
· Measure intermediate circuit voltage at D110.X2 and X1 (X2169/16)
· UZACT = 1,4 x Umains
o Caution! Direct voltage
· Actual value sensing D220 defective
· kV controller defective
· Tube arcing
· High-voltage generator H1 defective
· High-voltage cable OK?
· Filament current too high, perform tube adjust.