not booting

in display it will comes up to hd 11 xe after tha it will turn offs


   RE: Ultrasound Systems Report Post
Fri Mar 30 2012Reply from supperman
avatar placemark

HI

You should check power supply first. i have seen two times with this system.

 

Thanks


Report Post
Thu Jan 17 2013Reply from JerryW
avatar placemark

If the green LED over the on/off button dims or goes out it is the power supply.

I have changed 2 this month.


Report Post
Thu Jan 17 2013Reply from JerryW
avatar placemark

If the green LED over the on/off button dims or goes out it is the power supply.

I have changed 2 this month.


Report Post
Thu Jan 17 2013Reply from dferguson
dferguson

It is most likely a power supply issue; however, have seen the Display Processor Power Supply mimic different power supply issues.


 
 

Report Post
Sat Jan 19 2013Reply from sakher
avatar placemark

Its a problem with the power supply ,you have to change it 


Report Post