Hi, can help me with an issue of powersuite versapulse, at initial test flashlamp test failed, i had replaced 2 flashlamps for new ones but the fail is present anyway, the three flashlamps are good and flashing at initial test but just one is detected and the system is limitated to 20w.
Tue Nov 01 2016Reply from Zequn.Long
perhaps i can help you but i am not good in English.
all lamp can into  simmer state ? do you adjusted the total mirror or part mirror ?
Tue Nov 01 2016Reply from MSALAZAR
hi Zequn,
yes, all lamps are into simmer state, i dont move any adjust yet, can you tell me the process to adjust please?
Tue Nov 01 2016Reply from ChristineMae
ChristineMae
Adjusting the HR and OC are not an easy process. It would be ill advised to do so unless properly trained. Once the beam is lost it can be nearly impossible to find again. However If you would like some assistance in verifying the issue you may email me at info@umlasers.com and we can speak on how to put the laser into service mode.
Thank you,
Christine Ricoy
Christine Ricoy
C.L.R.T.