Just the gray keys below the display quit working on three different units from the same customer. Any ideas?
Softkeys not working
If you are in need of immediate assistance please feel free to reach out directly to our Defibrillator sponsors. They are happy to help.
Just be sure to tell them you were referred by MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
you could try to reseat the cables on the front panel membrane and the control knob pcb
If that doesn't repair the unit then its most likely the front panel membrane and or Control knob PCB failure.
were all the front panel membranes brand new?
Report Post
Reseating doesn't seem to help, at least not on the first one I opened up. They don't appear to be new. It just strikes me as odd that 3 membranes would die at the same time.
Quote ReNew14
you could try to reseat the cables on the front panel membrane and the control knob pcb
If that doesn't repair the unit then its most likely the front panel membrane and or Control knob PCB failure.
were all the front panel membranes brand new?
Report Post
You're telling me! I guess they could have abused them all the same way, but it still surprises me.
Quote ReNew14
that is very odd!?