Hi,
I have a Xario XG that fails to load/initiate the application post power up and returns with an error message "Start up is halted. Shutdown is recommended". Is there a solution for this? Any help would be great!
Thanks,
Noor
Hi,
I have a Xario XG that fails to load/initiate the application post power up and returns with an error message "Start up is halted. Shutdown is recommended". Is there a solution for this? Any help would be great!
Thanks,
Noor
If you are in need of immediate assistance please feel free to reach out directly to our Ultrasound System sponsors. They are happy to help.
Just be sure to tell them you were referred by MedWrench.
Your problem is with the software. reload or replace the HDD. It is probably a database corruption that has caused the issue
I cannot access the service mode from the 'system maintenance', the key stroke combination doesn't work. All the leds seem to be okay, there all green and lit (not flashing). I did aslo try to recreate the database from Windows environment but it returned with error message "E7031: UT_ER_DBID_NOACCESS: Failed to access databaseID file".
To Delete and re-create database
-Go into service - click menu then maintenace - hit ctrl+shift+next
-log in sceen displayed
-in service click command line
-type stopz.js click go (this will close imaging application)
-go to c:\setxg\aplioxgsetupmenu.exe
-a service menu is displayed
-click uncommon tools
-click remove database (all images will be gone) will take several min. wait for it to complete
-click create database this will also take several min. wait for it to complete
-go back to service tools page and click restart aplio
-reboot system and DB should be clear
-