ERROR CODE F56,F52

I need manual service for MUX 100H mobile art,,

can you send to my email, please

parsa.teb1@gmail.com

thank for your help me


   RE: Mobile Radiography Report Post
Sat Nov 19 2016Reply from bmet75
avatar placemark

Both of these errors are due to bad batteries. F52 is the battery voltage during charging is less then 100VDC. F56 is the negative battery pack is 12vdc higher then that positive pack. I would replace the batteries. You dont have to buy shimadzu batteries. We use to use batteries that were in UPS systems. 12V 9AH F2 connector.


Report Post
Tue Nov 22 2016Reply from parsa.teb
avatar placemark
Hello I charge the device at the time the error was F5F. Sw8-3 charge test mode is correct but after a minute error in the charging time ???? Please help me Thanks a lot

Report Post
Wed Dec 06 2023Reply from joseph fransua
avatar placemark

Tengo un error F56 que aparece después de haber reemplazado las baterías, ¿el error puede deberse a que una de las baterías vino defectuosa de fábrica?