No water pressure to assistant side

For an A-dec Cascade delivery system:

Have examined components but cannot determine what is causing the low pressure on the assistant side air/water syringe. I have ruled out the handpiece and any handpiece accessories. 40 PSI is coming out of floor box pressure gauge.


   RE: Power Report Post
Tue Nov 22 2016Reply from MedWrench MM
MedWrench MM

Is this a Cadwell Laboratories Cascade EMG system?


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Tue Nov 22 2016Reply from Celios
avatar placemark

This is an Adec Cascade dental delivery system. It's a very common model; not sure why only "200" is listed under Adec on MedWrench.


Report Post
Mon Mar 08 2021Reply from Chris-CBET
avatar placemark

did this ever get solved? I'd suggest checking the water filter on the floor reg, and cleaning/rebuilding the syringe. depending on water quality, i've found a lot of build up of corrosion inside those syringes


Report Post