Primary Audible alarm post

Self test pass

Goes into Biomed failure.

How to reset or repair

Primary Audible alarm Post.


   RE: Infusion Pump Report Post
Tue Dec 20 2016Reply from Hassan Serhan
avatar placemark

This is what the manual says about your System Failure Alarm "Primary Audible alarm Post".

Power-up self-test sensed no current flowing in the primary speaker. Speaker may be unplugged or wire came loose from connector. Check connections from main board to interconnect and from interconnect board to speaker. Replace speaker if wiring is intact.

Let me know if you need further help.

Hasen

(855) 477-8866


Report Post
Tue Dec 20 2016Reply from MedWrench MM
MedWrench MM

If you are in need of immediate assistance please feel free to reach out directly to our Infusion Pump sponsors. They are happy to help.

J2S Medical

RepairMED

Infusion Pump Repair

Elite Biomedical Solutions

 

Just be sure to tell them you were referred by MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Tue Dec 20 2016Reply from biomed6165
avatar placemark

Usually one of the capacitors on the IC board is faulty--often see a loose connection & sometimes it breaks off altogether.  Hope this helps.