I had concerns from users that they take the arterial blood from newborn baby for blood gas test, however they use the capillary. So I wonder that which types of measurement we must use (syringe mode or capillary mode)? And are there any problem if we use mix-up 2 modes?


   RE: Blood Gas/pH Analyzer Report Post
Wed Dec 21 2016Reply from Mc Grath Carlos
avatar placemark

Hi The diference is in the volume in one mode and other, and treatdment of the samply plus volume.

I dont think that is a best way to mixed it you can lose part of the sample test 


Report Post
Wed Dec 21 2016Reply from Merdo
avatar placemark
You must use Capillary mod please

Report Post
Wed Dec 21 2016Reply from Merdo
avatar placemark
But it is a diffucult work =