NIBP Issue

When testing the NIBP on Welch Allyn Propaq Encore Model 202 EL and Model 206 E the pump doesn't build up 200 mmHg. What could possible be the issue? Could it be the pump or just a setting in the setup of the monitor itself?


   RE: Vital Signs/Multi-Parameter Report Post
Mon Apr 09 2012Reply from GaryBiomed
GaryBiomed

I uploaded the service manual here

http://www.medwrench.com/?go.jump/no/5277

Page 40 has the NIBP cals.

From here you should be able to figure out if it's a setting or internal tubing or pump.