too hot in revelator bath

Hello everyone, I have a cp1000 agfa whose revelator bath is too hot and the film is burned.

I want the service manual thank you


   RE: Film Processor Report Post
Wed Jan 11 2017Reply from MedWrench MM
MedWrench MM

Hi Hayem117,

One of our users sent over the manual to me. You will now find it listed on the product page under the documents tab.

Hope this helps!


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Wed Mar 04 2020Reply from rtswapp
avatar placemark

THE WHAT? The Revelator bath?

No such thing. You mean the developer bath?

The thermostat is internal in the heating assembly. You need to replace the developer heater.


Report Post
Thu Mar 05 2020Reply from melanie b
avatar placemark

this worked for me:   find the logic boards under the drier assembly and film exit rack and swap out the fixative plug for the developer plug.   92 degrees fahrenheit is all you need for developer temperature.  The fixative does not need to be hot, leave it unplugged and at room temp.   If the wire is too short, extend it and put heat shrink tubing around the spliced in extention piece.   Think it thru, you can handle this.    I will try to post a picture of the logic board where the fixative and developer attach.