LL and RL leads

Got a machine, on a patient simulator, the unit displays LL disconnected..  Changed the LL cable, LL disconnected.  Changed the banana clip end, LL disconnected.  Changed the CAM module, LL disconnected.  Changed the CAM Cable, LL disconnected.  Put the RL on the LL, unit functions normally.  Thoughts?


   RE: Electrocardiograph (EKG / ECG) Report Post
Tue Jan 10 2017Reply from MedWrench MM
MedWrench MM

If you are in need of immediate assistance please feel free to reach out directly to our Electrocardiograph (EKG/ECG) sponsors. They are always happy to help out when they can.

AIM

Southeastern Biomedical

 

Just be sure to tell them you were referred by MedWrench.


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Tue Jan 10 2017Reply from jroberts
jroberts

When you say "changed the CAM Cable", do you mean the cable running from the CAM  to the MAC 5500.  If so, I suggest replacing or repairing the CAM board itself. 


-J
Report Post
Tue Jan 10 2017Reply from Crutch
avatar placemark

I will try a known good cam module and see if perhaps that is the problem.  I have tried different cam modules, all with the same indications, so maybe (since they are on my desk) they are all broken too.

 

Thank you


Report Post
Tue Jan 10 2017Reply from Crutch
avatar placemark

I am happy to report that the problem has been resolved.  It appeared that no matter which CAM module I used, or how I hooked it up it always failed the same way.  Turns out there is a problem with my patient simulator.  I used a different simulator and had no problems with the unit.

 

I hadn't thought about test equipment being faulty.