Does anybody know what an error 166 is on a Valleylab Ligasure?
Tue Jan 10 2017
Error 166
Tue Jan 10 2017Reply from MedWrench MM
Hi Moderlin,
If you are in need of immediate assistance please feel free to reach out directly to our Electrosurgical Unit (ESU) sponsors. They are happy to help out when they can.
Just be sure to tell them you were referred by MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Tue Jan 10 2017Reply from Bill McSharry
The service manual says:
DOSERR_ECON_HV_ERROR
Sys_Econ and HV_SNS do not compare within limits during keying.
Turn the power switch off and on again.
If the error reappears, replace the control board.
If the error reappears, replace the high voltage power supply board.
Report Post