Fri Apr 13 2012
Preventive Maintenance Required
How do you remove the "preventive maintenace required" from the monitor screen on the Dash 3000
Sat Apr 14 2012Reply from
I don't remember the model we have but if they are the same the message comes up for NIBP calibration.  Do the calibration the service manual shows and then get out of the service menue and go back to patient monitoring mode.  Now discharge the patient and then reboot (off/on) the monitor.  The message should be gone.
Mon Apr 16 2012Reply from lloyd
Discharged the the admitted patients. Go to admit window and discharged. Again if still the message appear do the same procedure until the message disappear.
Hope this will help you
Mon Apr 16 2012Reply from landa2020
Thank you very much for both of the above answers.   You are both correct and this took care of the problem.   Thank you for spending the time to help.
