Fluid levels too low...

AGFA CP 1000 film processor. Fluid level in all tray(tanks) too low. Replaced all 3 valves on replenisher bottles, no difference. Two different techs came out to fix, but couldn't. Tanks will fill when "manual replenishment" switch is activated - but this shouldn't be necessary. Any ideas??? Jerry


   RE: Film Processor Report Post
Fri Apr 20 2012Reply from MedWrenchManager
MedWrenchManager

Check processor level. Very important on this model. Also check the drain tubes that empty the trays fluid may be slowly leaking by have someone replace all 3 tray drain tubes.

Answer provided by Raymond M,, Sales manager at Superior Radiographics