Thu Feb 09 2017

faulty d925 board

Seimens D925 faulty electronic board

   RE: Mobile Radiography Report Post
Thu Feb 09 2017Reply from Ayazra
avatar placemark
there is a burned out 8 legs IC on the board it is needed to now the number ID of it to repair

Report Post
Thu Feb 09 2017Reply from MedWrench MM
MedWrench MM

D925 is the model number for the electronic board. What equipment will this be used with?


Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Report Post
Sat Feb 11 2017Reply from Ayazra
avatar placemark
It is used in Seimens Polymobil III Plus ( Mobile X-Ray Equipment ). Thanks & Best Regards 11/02/2017 Eng. Ayazra