Co2 values are not comming

Help! The Co2 values are not comming. Waveforms are present but values are not appearing on the display.


   RE: Medical Gas Monitor Report Post
Tue Apr 24 2012Reply from MedWrenchManager
MedWrenchManager

 Replace pump tubing part of annual pm.

Answer provided by Tracy W., Lead Equipment Specialist at 59th MRLS Clinical Engineering Flight


Report Post
Mon Apr 30 2012Reply from MedWrenchManager
MedWrenchManager

Also be sure to perform a gas calibration procedure after you replace the tubing.

Answer provided by Mark S., Senior Biomedical Equipment Technician at North Mississippi Medical Center


Report Post
Thu Oct 11 2012Reply from MedWrenchManager
MedWrenchManager

The pump should be at 180 - 200 ml / hr. if memory serves. A new PM kit couldn't hurt either. Good Luck.... 

Answer provided by Gregorgy W., Biomedical Engineering Dept Manager at Reonovo Solutions via LinkedIN.


Report Post