Head section raises but wont lower
Thu Feb 23 2017
totalcare P1900
Fri Feb 24 2017Reply from MedWrench MM
Try reaching out directly to our Bed sponsors. They are happy to help out when they can.
MedMattress
DiaMedical USA
Just be sure to tell them you were referred by Jonathan at MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Tue Feb 28 2017Reply from Joe Sosnowski
Thu Mar 02 2017Reply from kle257
The Head Down valve and solenoid is S8 on the hydraulic block.
Does the head go down when you engage the CPR function? That lets me know if the issue is with the cylinder or the valve/solenoid.
Check the voltage to the S8 solenoid when the function switch is pressed to rule out switch and wiring issues.
Push the function switch and touch the top of the valve with the tip of a screw driver. Is it magnetized? If not it could be a faulty solenoid.
Prop the head section (so it does not drift down on you) and remove the valve/solenoid? Does fluid leak? Could be a blockage in the hydraulic block. There is a screen in there.
If fluid flows, replace the valve. Don't forget to fish out the old o-ring.
Good Luck...
Sun Jul 18 2021Reply from Crazy8dot