I have replaced the print-head assembly and during self test the "R" wave prints with a continuous fade. Has anyone come across the same issue?
Fri Mar 10 2017
Faded "R" wave while printing
Wed Mar 15 2017Reply from MedWrench MM
If you are in need of immediate assistance please feel free to reach out directly to our Electrocardiograph (EKG/ECG) sponsors. They are happy to help out when they can.
Just be sure to tell them you were referred by Jonathan at MedWrench.
Jonathan F. Payne,
Media Manager @ MedWrench,
866-989-7057,
Jpayne@MedWrench.com
Wed Mar 15 2017Reply from Robert Quick, CBET, USAF Ret.
Change the battery! I know this sounds weird for a print issue, but according to GE the battery effects the print. We have had this issue a few times in the last year and changing thbattery cleared it up each time. Good luck!
Robert A. Quick, BS, CBET, TSgt. (USAF Ret.)
Biomedical Equipment Technician
University of Tennessee Medical Center
(865) 305-4607 RQuick@UTMCK.edu
Fri Mar 31 2017Reply from BMET32